Seeking tips on using J2121A High Power Line Injector to measure boost PFC inductors inductance with bias currents up to 20Apk and in the frequency range of 100kHz-500kHz. From what I have found there is not much published on using this particular injector in this manor and what needs to be done to make the measurements as accurate as possible. Also looking for information on if the current limit (20A) and voltage limits (400V) can be slightly exceeded without damage when measuring converter input & output impedances. As it is common for PFC output voltages / Isolation stage input voltages to be in the 390-450V range.
top of page
bottom of page
yes, please email me at Steve@picotest.com
I don't know how much higher you want to go in current, but I believe you can probably push it to 450V. It would exceed our derating guidelines, but the capacitors are rated for 500V. At some point you'll thermally kill it, but if you wanted to go to 25Amp I would say maybe, especially if you don't run it there continuously. As a controlled experiment, I think it would be ok...
We do have some application info and also did some videos and webinars I can send, but they won't be all that helpful for you. We typically show step down regulator inductors, which are much lower value, since we don't get many requests for higher inductances. At high inductance, you will run into the noise limitations of the internal hall sensor as you have seen.
The voltage drop of the J2121A is only 700mV, so realistically, you can't modulate more than about 500mVpk, that's why increasing the signal level doesn't really help at some point.
I think you have 4 options to improve your measurements.
Try reducing the RBW from 100 Hz to 10Hz
Try trace averaging, which I admit, I haven't used myself
Use an external current probe. That will be lower noise than the internal hall sensor
Use a current sense resistor
This is a great application, so we'll help as much as you need and maybe end up with an app note for PFC inductors in the end. Can you send us an inductor we can try to measure here? What is the bias current?
We have two methods that will work for you. Here is an app note for the J2121A and I also did a presentation at PSMA2020. You can find the video presentation here https://www.dropbox.com/s/0485cltxclli3ws/PSMA%202020.mp4?dl=0